LSC vs GM for smear release

And after a short P‑32 radiolabeling run (about 5 µCi, 20 min at the hood), my GM pancake at about 3 cm gives a clean 10 s scan on bench-top smears, yet LSC on the same wipes in [redacted] Ultima Gold reads about 25 cpm above background once I apply the quench curve… In practice, do you let the LSC result drive the release decision for the area, or do you set a dual criterion, and how are you documenting that for radiation safety sign-off? I tend to defer to the more sensitive radiochemical method, but I want our RPP to be consistent and defensible.

‌⁠‍⁠​‍​‍‌⁠‌​​‍​‍​⁠‍‍​‍​‍‌‍​⁠‌‍‍​‌‍‌‌‌‍⁠‌‌‍‍‌‌⁠​⁠‌⁠‌​​‍​‍​‍⁠​​‍​‍‌‍‍⁠​‍​‍​⁠‍‍​‍​‍‌‍⁠‍‌‍‌‌‌⁠‌⁠‌‌⁠⁠‌⁠‌​‌‍⁠⁠‌⁠​​‌‍‍‌‌‍​⁠​‍​‍​‍⁠​​‍​‍‌‍‍‌‌‍‌​​‍​‍​⁠‍‍​‍​‍‌‍⁠‍‌‍‌‌‌⁠‌⁠​‍​‍​‍⁠​​‍​‍‌‍‌​​‍​‍​⁠‍‍​‍​‍​⁠​‍​⁠​​​⁠​‍​⁠‌‌​⁠​‌​⁠​‌​⁠​​​⁠‌‍​‍​‍​‍⁠​​‍​‍‌‍‍​​‍​‍​⁠‍‍​‍​‍‌‍‌‌‌‌​⁠‌​‍⁠‌⁠​‌‌‍⁠‌‌‍⁠​​‍⁠‌‌​‌​‌​‍‌‌⁠‍‍‌​‍‍‌​​⁠‌‍‌⁠‌⁠​​‌​‌⁠​⁠‌⁠​‍​‍‌⁠⁠‌​

, LSC flags tiny bumps while the pancake is quiet. > quench curve… In practice, do you let the LSC result drive the release decision — for P-32 we go dual: if the GM sweep at about 3 cm is clean and LSC is at or below its MDA, we release; about 25 cpm over bg wouldn’t stop us. Caveat: validate the quench curve with a P-32 spiked wipe and keep wipes nearly dry so solvent/cocktail doesn’t skew it.

‌⁠‍⁠​‍​‍‌⁠‌​​‍​‍​⁠‍‍​‍​‍‌‍​⁠‌‍‍​‌‍‌‌‌‍⁠‌‌‍‍‌‌⁠​⁠‌⁠‌​​‍​‍​‍⁠​​‍​‍‌‍‍⁠​‍​‍​⁠‍‍​‍​‍‌⁠​‍‌‍‌‌‌⁠​​‌‍⁠​‌⁠‍‌​‍​‍​‍⁠​​‍​‍‌‍‍‌‌‍‌​​‍​‍​⁠‍‍​⁠‌‌​⁠​​​⁠‍​​⁠​⁠​‍⁠​​‍​‍‌‍‌​​‍​‍​⁠‍‍​‍​‍​⁠​‍​⁠​​​⁠​‍​⁠‌‌​⁠​‌​⁠​‌​⁠​​​⁠‍‌​‍​‍​‍⁠​​‍​‍‌‍‍​​‍​‍​⁠‍‍​‍​‍‌​​‍‌⁠‌‌‌⁠‌‍‌⁠‍​‌‍⁠‌​⁠‍​‌​‌⁠‌⁠​‌‌⁠‍‌‌⁠‍​‌‌‍​‌​‍‌‌⁠‌‌‌​‍⁠‌​‌​​⁠‌​​‍​‍‌⁠⁠‌